Skip to content
  • Categories
  • Recent
  • Tags
  • Popular
  • World
  • Users
  • Groups
Skins
  • Light
  • Cerulean
  • Cosmo
  • Flatly
  • Journal
  • Litera
  • Lumen
  • Lux
  • Materia
  • Minty
  • Morph
  • Pulse
  • Sandstone
  • Simplex
  • Sketchy
  • Spacelab
  • United
  • Yeti
  • Zephyr
  • Dark
  • Cyborg
  • Darkly
  • Quartz
  • Slate
  • Solar
  • Superhero
  • Vapor

  • Default (No Skin)
  • No Skin
Collapse
Code Project
  1. Home
  2. The Lounge
  3. Measuring your multimeter while attempting to measure your circuit

Measuring your multimeter while attempting to measure your circuit

Scheduled Pinned Locked Moved The Lounge
questioncsscomdata-structureslearning
40 Posts 15 Posters 0 Views 1 Watching
  • Oldest to Newest
  • Newest to Oldest
  • Most Votes
Reply
  • Reply as topic
Log in to reply
This topic has been deleted. Only users with topic management privileges can see it.
  • R raddevus

    I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

    3V / 10 Ohm = 0.3A (300mA)

    I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

    157.7 mA

    What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

    3V / 0.157 = 19.10 (Ohms)

    So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

    Richard Andrew x64R Offline
    Richard Andrew x64R Offline
    Richard Andrew x64
    wrote on last edited by
    #3

    Maybe you need a more accurate meter: Fluke[^] :)

    The difficult we do right away... ...the impossible takes slightly longer.

    1 Reply Last reply
    0
    • R raddevus

      I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

      3V / 10 Ohm = 0.3A (300mA)

      I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

      157.7 mA

      What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

      3V / 0.157 = 19.10 (Ohms)

      So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

      L Offline
      L Offline
      Lost User
      wrote on last edited by
      #4

      I'm weighing 500 gr of rice. I put the scale to 0, and put the bowl on the scale. What's that, it weighs 153 grams? It's called calibrating, and your measuring-device should know how to compensate for its own interference.

      Bastard Programmer from Hell :suss: If you can't read my code, try converting it here[^]

      G 1 Reply Last reply
      0
      • R raddevus

        I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

        3V / 10 Ohm = 0.3A (300mA)

        I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

        157.7 mA

        What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

        3V / 0.157 = 19.10 (Ohms)

        So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

        L Offline
        L Offline
        Lost User
        wrote on last edited by
        #5

        And how is about to replace your 10 Ohm resistor by e.g. a 100 Ohm? Can you confirm with 100 Ohm your Observation?

        It does not solve my Problem, but it answers my question

        R 1 Reply Last reply
        0
        • R raddevus

          I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

          3V / 10 Ohm = 0.3A (300mA)

          I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

          157.7 mA

          What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

          3V / 0.157 = 19.10 (Ohms)

          So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

          C Offline
          C Offline
          CodeWraith
          wrote on last edited by
          #6

          Don't forget the tolerance of your resistors. Look at the last color band. Tolerances of 10% or worse were not uncommon, but today resistors with low tolerances are not so expensive anymore.

          I have lived with several Zen masters - all of them were cats. His last invention was an evil Lasagna. It didn't kill anyone, and it actually tasted pretty good.

          R 1 Reply Last reply
          0
          • R raddevus

            I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

            3V / 10 Ohm = 0.3A (300mA)

            I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

            157.7 mA

            What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

            3V / 0.157 = 19.10 (Ohms)

            So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

            G Offline
            G Offline
            GuyThiebaut
            wrote on last edited by
            #7

            It's no big deal, it's just a case of knowing the error bars for your measurement. Every measurement in science is meaningless without knowing the precision of the measuring instrument.

            “That which can be asserted without evidence, can be dismissed without evidence.”

            ― Christopher Hitchens

            C 1 Reply Last reply
            0
            • L Lost User

              I'm weighing 500 gr of rice. I put the scale to 0, and put the bowl on the scale. What's that, it weighs 153 grams? It's called calibrating, and your measuring-device should know how to compensate for its own interference.

              Bastard Programmer from Hell :suss: If you can't read my code, try converting it here[^]

              G Offline
              G Offline
              GuyThiebaut
              wrote on last edited by
              #8

              You also need to know the sensitivity of measurement. So while you may have zeroed your scale you will also need to know that the scale is, for example, accurate to within 1 gram.

              “That which can be asserted without evidence, can be dismissed without evidence.”

              ― Christopher Hitchens

              L 1 Reply Last reply
              0
              • G GuyThiebaut

                You also need to know the sensitivity of measurement. So while you may have zeroed your scale you will also need to know that the scale is, for example, accurate to within 1 gram.

                “That which can be asserted without evidence, can be dismissed without evidence.”

                ― Christopher Hitchens

                L Offline
                L Offline
                Lost User
                wrote on last edited by
                #9

                I'd assume the last digit on the scale to be a guess. That's still a lot less than the weight of the scale itself :)

                Bastard Programmer from Hell :suss: If you can't read my code, try converting it here[^]

                1 Reply Last reply
                0
                • G GuyThiebaut

                  It's no big deal, it's just a case of knowing the error bars for your measurement. Every measurement in science is meaningless without knowing the precision of the measuring instrument.

                  “That which can be asserted without evidence, can be dismissed without evidence.”

                  ― Christopher Hitchens

                  C Offline
                  C Offline
                  CodeWraith
                  wrote on last edited by
                  #10

                  AND the tolerances of the parts you are measuring, in this case the resistor.

                  I have lived with several Zen masters - all of them were cats. His last invention was an evil Lasagna. It didn't kill anyone, and it actually tasted pretty good.

                  1 Reply Last reply
                  0
                  • R raddevus

                    I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

                    3V / 10 Ohm = 0.3A (300mA)

                    I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

                    157.7 mA

                    What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

                    3V / 0.157 = 19.10 (Ohms)

                    So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

                    R Offline
                    R Offline
                    Roger Wright
                    wrote on last edited by
                    #11

                    There are several factors affecting the measurement; the meter most certainly does not have an internal impedance anything close to 10 ohms. Most likely is that you're far overloading the battery - 300 mA is a lot for a button cell or even a small alkaline battery. They're designed for a discharge rate much lower, and their internal resistance increases when you try to pull more current than they can deliver. Even a cheap multimeter is expected to measure 10A without disturbing the circuit - perhaps a 100 mV drop = .01 ohm. Try again using a larger resistor.

                    Will Rogers never met me.

                    R 1 Reply Last reply
                    0
                    • R raddevus

                      I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

                      3V / 10 Ohm = 0.3A (300mA)

                      I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

                      157.7 mA

                      What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

                      3V / 0.157 = 19.10 (Ohms)

                      So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

                      CPalliniC Offline
                      CPalliniC Offline
                      CPallini
                      wrote on last edited by
                      #12

                      Quote:

                      So it looks like my meter has around 9-10 Ohms

                      That's difficult to believe. A quick and dirty test with my cheap multimeter gave a internal resistance of about 0.2 Ohms while measuring about 0.1 A (2.2 V lab power supply on 22 Ohm resistor). By the way, is the resistor properly sized (common 1/4W ones cannot stand the 0.9 W of your test)?

                      In testa che avete, signor di Ceprano?

                      R 1 Reply Last reply
                      0
                      • J Jorgen Andersson

                        Don't forget the internal resistance of the battery or power source. 9-10 Ohms sounds like a bit much for the multimeter.

                        Wrong is evil and must be defeated. - Jeff Ello

                        R Offline
                        R Offline
                        raddevus
                        wrote on last edited by
                        #13

                        I watched a YouTube that explains it very well #194: What is ammeter burden voltage, and why you should care. - YouTube[^] And someone else pointed out a special port on my meter (marked 20A) which has a smaller burden and I was was able to read 270mA on the circuit. Interesting that it was the meter and not the battery.

                        J 1 Reply Last reply
                        0
                        • L Lost User

                          And how is about to replace your 10 Ohm resistor by e.g. a 100 Ohm? Can you confirm with 100 Ohm your Observation?

                          It does not solve my Problem, but it answers my question

                          R Offline
                          R Offline
                          raddevus
                          wrote on last edited by
                          #14

                          I think that is a great idea and will try that and let you know. I also learned a bit more about this see : This thread the Lounge[^]

                          L 1 Reply Last reply
                          0
                          • C CodeWraith

                            Don't forget the tolerance of your resistors. Look at the last color band. Tolerances of 10% or worse were not uncommon, but today resistors with low tolerances are not so expensive anymore.

                            I have lived with several Zen masters - all of them were cats. His last invention was an evil Lasagna. It didn't kill anyone, and it actually tasted pretty good.

                            R Offline
                            R Offline
                            raddevus
                            wrote on last edited by
                            #15

                            Good point and I did measure resistor on meter a d it gets 9.8 to 10.:thumbsup:

                            1 Reply Last reply
                            0
                            • R raddevus

                              I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

                              3V / 10 Ohm = 0.3A (300mA)

                              I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

                              157.7 mA

                              What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

                              3V / 0.157 = 19.10 (Ohms)

                              So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

                              CPalliniC Offline
                              CPalliniC Offline
                              CPallini
                              wrote on last edited by
                              #16

                              By the way, did you check the battery voltage when it was actually connected to the resistor?

                              In testa che avete, signor di Ceprano?

                              R 1 Reply Last reply
                              0
                              • R Roger Wright

                                There are several factors affecting the measurement; the meter most certainly does not have an internal impedance anything close to 10 ohms. Most likely is that you're far overloading the battery - 300 mA is a lot for a button cell or even a small alkaline battery. They're designed for a discharge rate much lower, and their internal resistance increases when you try to pull more current than they can deliver. Even a cheap multimeter is expected to measure 10A without disturbing the circuit - perhaps a 100 mV drop = .01 ohm. Try again using a larger resistor.

                                Will Rogers never met me.

                                R Offline
                                R Offline
                                raddevus
                                wrote on last edited by
                                #17

                                That's a great point and helps me to learn. However I am using 2 AA in series to obtain 3v. I still thought 300mA might be a bit too much to pull but in the spirit of experimentation I rolled on. :)

                                R 1 Reply Last reply
                                0
                                • CPalliniC CPallini

                                  Quote:

                                  So it looks like my meter has around 9-10 Ohms

                                  That's difficult to believe. A quick and dirty test with my cheap multimeter gave a internal resistance of about 0.2 Ohms while measuring about 0.1 A (2.2 V lab power supply on 22 Ohm resistor). By the way, is the resistor properly sized (common 1/4W ones cannot stand the 0.9 W of your test)?

                                  R Offline
                                  R Offline
                                  raddevus
                                  wrote on last edited by
                                  #18

                                  Here's a very good short YouTube which I was guided to that explains it and the one meter the presenter uses does indeed have 10 ohms of resistance. #194: What is ammeter burden voltage, and why you should care. - YouTube[^] Also I was guided to use another port on my meter (marked 20A) which has more like 1 ohm resistance and I was able to read 270mA.

                                  1 Reply Last reply
                                  0
                                  • CPalliniC CPallini

                                    By the way, did you check the battery voltage when it was actually connected to the resistor?

                                    R Offline
                                    R Offline
                                    raddevus
                                    wrote on last edited by
                                    #19

                                    I measure approx 2.2 volts across the 10 Ohm resistor in my circuit. I'm assuming that would mean something like:

                                    2.2 / 10 Ohm = 220mA

                                    However, I know when I measure that I only see 157mA of current. So...

                                    220mA - 157mA = 63mA (which are missing somewhere).

                                    I plug in the missing 63 like so:

                                    .8V (voltage drop) / .063 = 12.69 Ohms

                                    So I'm still seeing a extra ~12 Ohms that I can't account for. That is obviously higher than my 9-10 Ohm extra I was calculating but could be due to rounding. Does that seem at least somewhat correct? I'm just noodling my way through this to learn more. Thanks,

                                    L CPalliniC 2 Replies Last reply
                                    0
                                    • R raddevus

                                      That's a great point and helps me to learn. However I am using 2 AA in series to obtain 3v. I still thought 300mA might be a bit too much to pull but in the spirit of experimentation I rolled on. :)

                                      R Offline
                                      R Offline
                                      Roger Wright
                                      wrote on last edited by
                                      #20

                                      Carry on! :-D Electronics remains my first love, despite nearly 40 years since school! It's fascinating, always...

                                      Will Rogers never met me.

                                      1 Reply Last reply
                                      0
                                      • R raddevus

                                        I measure approx 2.2 volts across the 10 Ohm resistor in my circuit. I'm assuming that would mean something like:

                                        2.2 / 10 Ohm = 220mA

                                        However, I know when I measure that I only see 157mA of current. So...

                                        220mA - 157mA = 63mA (which are missing somewhere).

                                        I plug in the missing 63 like so:

                                        .8V (voltage drop) / .063 = 12.69 Ohms

                                        So I'm still seeing a extra ~12 Ohms that I can't account for. That is obviously higher than my 9-10 Ohm extra I was calculating but could be due to rounding. Does that seem at least somewhat correct? I'm just noodling my way through this to learn more. Thanks,

                                        L Offline
                                        L Offline
                                        Luc Pattyn
                                        wrote on last edited by
                                        #21

                                        Combining a cheap multimeter and an overloaded battery may lead to all kinds of silly conclusions. If you want to comment on a multimeter, make sure you use a lab-grade power supply and resistor both capable to act nominally in your circuit; and likewise when checking a battery, make sure to use a lab-grade multimeter and resistor. And make sure none of the components involved heats up (which indicates operation outside the nominal operating range), as currently your resistor probably and your battery most likely does. :)

                                        Luc Pattyn [My Articles] Nil Volentibus Arduum

                                        1 Reply Last reply
                                        0
                                        • R raddevus

                                          I'm working on Chapter 4 of Practical Electronics For Makers and I'm working the reader through some very simple circuits so we can see how resistors affect (lower) current. So I set up the simplest circuit that is simply a 3V battery supply and one 10 Ohm resistor. Fist I calculate the expected current value I will see using Ohm's law (E = IR).

                                          3V / 10 Ohm = 0.3A (300mA)

                                          I hook up the circuit to run through my multimeter so I can measure (and confirm) the value. My meter displays:

                                          157.7 mA

                                          What? Google... Apparently meters have their own internal resistance. I decide to turn Ohm's law around and calculate internal resistance.

                                          3V / 0.157 = 19.10 (Ohms)

                                          So it looks like my meter has around 9-10 Ohms of its own resistance. Hmmm... Measuring the measurements of a measuring device.:~ Of, course if I had higher resistance in my circuit I'd probably not notice this, because it is only 10 Ohms. I posted a question about this on electronics stack exchange where you can see pictures and the circuit. Multimeter, measuring current calculates less. Can I calculate meter's internal resistance this way? - Electrical Engineering Stack Exchange[^]

                                          P Offline
                                          P Offline
                                          Peter_in_2780
                                          wrote on last edited by
                                          #22

                                          In addition to the other sources of error (I'd wager battery internal resistance is the biggest), you've run into ammeter burden[^]. In most modern digital multimeters, the maximum burden is approximately the full scale on your lowest voltage range, probably something like 200mV. (It measures current by the voltage drop across a shunt.) Welcome to real-world electronic engineering! Cheers, Peter

                                          Software rusts. Simon Stephenson, ca 1994. So does this signature. me, 2012

                                          R 1 Reply Last reply
                                          0
                                          Reply
                                          • Reply as topic
                                          Log in to reply
                                          • Oldest to Newest
                                          • Newest to Oldest
                                          • Most Votes


                                          • Login

                                          • Don't have an account? Register

                                          • Login or register to search.
                                          • First post
                                            Last post
                                          0
                                          • Categories
                                          • Recent
                                          • Tags
                                          • Popular
                                          • World
                                          • Users
                                          • Groups